True enough, but the problem here is not a wavey baseline or steep baseline slope.What I meant is that it does not show that much the wavey baseline you see with pump-mixed eluents.
The real problem is inadequate sensitivity caused by the mobile phase background light absorption.
As I implied earlier if the mobile phase was transparent for light at 195 nm you could inject say 10 times less sample solution which would result in a symmetrical peak as well as narrow one. That’s my point.
My advice: remove the TFA. If the pH needs to be low, use phosphoric acid or better yet phosphate buffer. Then inject 10 – 20 µL instead of 100.
Regarding validation I understand you rationales but unfortunately still disagree. In my world validation means documentation of a working method - i.e. producing reliable results.
Best Regards
Best Regards
Dancho Dikov