The column length is 300mm and the guard column is 1/6th of column length i.e. 50 mm.
Both are made of the same material.
Does your analyte have a retention time of about 18 minutes without using a guard column?
Check the back pressure of 1.) system without any column, 2.) column without guard column, 3.) column with guard column and 4.) guard column without column. Post the values here.
1. System without any column- 131 psi
2. Column without guard column- 537 psi
3. Column with guard column- 682 psi
4. Guard column without column- 334 psi
1: Backpressure from System = 131 psi
2: Backpressure from Column = 537 -131 = 406 psi
4: Backpressure from Guard column = 334 - 131 = 203 psi
Therefore the calculated pressure with column and guard column should be theoretically: 131 + 406 + 203 = 740 psi
Your measured value 3) = 682 psi is about 9 % below of the calculated backpressure. In absolute terms of (740 - 682) 58 psi difference, this difference from the calculated value (about 4 bar absolute) is absolutely negligible.