Re: Problem with detecting last 3 PAHs
Posted: Fri Mar 07, 2014 9:35 am
Hi,
I am using Restek liners because our local Thermo representative will not work with me because I have an old Thermo device bought second-hand. The only ones to help me install the device and sell me consumables were the Agilent representatives.
So, this one I am using, is a Restek Open Top Uniliner, 4mm x 5.4 x 79.5 for Thermo GCs.
Yes, i believe my column is too low in the injector, since the tech guy that helped me to install the device and connected my column was an Agilent employee and set it up for 2mm in injector. I think i should stick it into the liner's neck. Now is exactly how I draw it in the first picture.
I am using exactly 3 seconds pre-injection and 2 seconds post injection dwell time.
This column setup might be a part of the problem. Other part might be due to "no pulse mode" I employ. It seems everybody use this pulse mode, but I can't. What I can do is loose the constant velocity of 40 cm/s I am using now, and make a pressure program. I can try to initially set up a bigger pressure, to force more vapors into the column, then make a ramp calculated to match the old 40 cm/s program based on column dimensions and temperature.
Monday I will start again the device and give a try with all these. Till then I am gathering ideas and tips
Thank you,
Vlad
I am using Restek liners because our local Thermo representative will not work with me because I have an old Thermo device bought second-hand. The only ones to help me install the device and sell me consumables were the Agilent representatives.
So, this one I am using, is a Restek Open Top Uniliner, 4mm x 5.4 x 79.5 for Thermo GCs.
Yes, i believe my column is too low in the injector, since the tech guy that helped me to install the device and connected my column was an Agilent employee and set it up for 2mm in injector. I think i should stick it into the liner's neck. Now is exactly how I draw it in the first picture.
I am using exactly 3 seconds pre-injection and 2 seconds post injection dwell time.
This column setup might be a part of the problem. Other part might be due to "no pulse mode" I employ. It seems everybody use this pulse mode, but I can't. What I can do is loose the constant velocity of 40 cm/s I am using now, and make a pressure program. I can try to initially set up a bigger pressure, to force more vapors into the column, then make a ramp calculated to match the old 40 cm/s program based on column dimensions and temperature.
Monday I will start again the device and give a try with all these. Till then I am gathering ideas and tips
Thank you,
Vlad