I'm using an Hypersil SAX column with 0.33g/l phtalic acid, brought at pH 6.4.
Column temperature is 35°C.
When I injected yesterday morning my first reference, retention time was about 11.5 minutes.
After injecting 9 references (Na2SO4 in water), the retention time was already decreased to 9.7 minutes.
After several hours, the RT was about 6 minutes, and peaks are getting distorted/unseparated from other peaks in my sample.
Any ideas about this decrease, or how to get my retention back?
Could you do a test run that came with the column? If the retention times of those standard anions are significantly lower, it means that the anion exchanger has lost some capacity. Sulfate ion (or divalents) are very sensitive to capacity changes on the stationary phase and any changes in the mobile phase composition (including CO2 contamination).
Secondly, are you using the recommended mobile phase? I am not sure that a totally aqueous phase is recommended on a silica substrate. There is decent work from the Carr's group which shows that the major metallic contamination comes from the metallic frits rather than the column body (or the pumps)-so passivating might not work in your case.
The suggestion to use polymeric substrates comes from the fact the polymers can be subjected to harsh cleaning steps with acids and bases - these reagents would simply chew up a silica column.