What kind of pump are you using? (metal-free?)
Advertisement
Discussions about IC and related topics
Could you do a test run that came with the column? If the retention times of those standard anions are significantly lower, it means that the anion exchanger has lost some capacity. Sulfate ion (or divalents) are very sensitive to capacity changes on the stationary phase and any changes in the mobile phase composition (including CO2 contamination).I'm using an Hypersil SAX column with 0.33g/l phtalic acid, brought at pH 6.4.
Column temperature is 35°C.
When I injected yesterday morning my first reference, retention time was about 11.5 minutes.
After injecting 9 references (Na2SO4 in water), the retention time was already decreased to 9.7 minutes.
After several hours, the RT was about 6 minutes, and peaks are getting distorted/unseparated from other peaks in my sample.
Any ideas about this decrease, or how to get my retention back?
Polymeric SAX are not "immune" from collecting transition metals, however, they don't trap them as notoriously as silica-based columns. For inorganic ion analysis, I'd tend to think polymeric SAX are more popular (by far?) compared to silica-based supports. Dionex, Metrohm, Hamilton-- all manufacture PS-DVB SAX.Is a polymeric SAX column "resistant" against transition metals?
We have 14 HPLC, running all reversed phase chromatography, it is a bit impractical to passivate them one by one, just for 1 application.
Would adding EDTA to the mobile phase be sufficient?