The last possibilities are the Active Inlet Valve or the Pump Seals?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
"Some months ago"? That can be a long timeThe active inlet valve is OK, because it's cartridge was changed some months ago. The outlet ball valve looks nice too.[/img]
Sorry, but I don't know why picture didn't attach.There is no screenshot.
Which solvents/column do you use?
Thank you for advice!"Some months ago"? That can be a long time.
"Looks nice"? Look is one thing, does it WORK nicely?
With pressure fluctuations, check valves are my first suspects. Don't trust them, check them!
Thanks, I shall try it.Hello,
I'll suggest a little test-method. 5min 100% solvent A, 5 min 100% A -> 100% B, 5min 100% solvent B.
Then you should see, which channel is the problem.