23 posts Page 2 of 2
I ran the system with the degaser OFF. Please see links for the baselines with and w/o flow.

https://docs.google.com/file/d/0Bw8z-jp ... sp=sharing
https://docs.google.com/file/d/0Bw8z-jp ... sp=sharing

The 20-min cycles remain after I turned the degaser off. But the cycles disappear when the flow is zero. Could it be the pump, the check valve that causes the cyclic noise?
You're making progress! Next question, are you pre-mixing your mobile phase or mixing on-line? If you're mixing on-line, it could be a proportioning problem. To diagnose, try premixing your mobile phase and put all inlet lines in the same reservoir.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Do you also record pressure?
Maybe there is a relationship between pressure and noise?

Ace
There is no solvent mixing issue. The mobile phases used in the last two chromatograms are water.

Could it be a faulty pump? How to record pressure in a Waters 2695 system?
In your instrument method.
The tab 2690/5 > Channel > enable channel.

Ace
How are you controlling your column temperature ? The 25 C that you give is very close to default room temperature, and this can make it difficult for a column oven not to overshoot its set point. What happens if you increase column temperature to say 30C. If you do not have a column oven you may be seeing the effects of fluctuating room temperature (although I would be surprised if they were so regular.

Peter
Peter Apps
When troubleshooting a system, do remove the column, that makes it easier to find out what hardware if any is not functioning properly.
With column OFF, place instaed a long piece of very narrrow ID tubing to create some back pressure. Then run your baseline. If you do NOT get a good baseline, then your problem probably is , A: an inlet check valve failing, B: solvents need degassing, or [ and I have seen this], a broken plunger.
THIS is 905 of most failing baseline stability ssiues.
Dear Forum:

Please help me with the baseline problem. I set up Waters 2695 Alliance with a 2998 PDA detector using the following parameter:

Column: Whatman Partsil 10-SCX
Column T: 25 celcius degree
Wavelength: 210 nm
Mobile phase: 25% 25 mM KH2PO4 adjusted to ph=4.5; 75% acentonitrile

After I flushed the system (with column and detctor flow cell) with water and 100% methanol, I still couldn't get a stable baseline.
Here is what the baseline looks like:
https://docs.google.com/file/d/0Bw8z-jp ... sp=sharing

Please help trouble shooting the baseline.

Thanks so much!
C.Tony Vella Royal British Legion
WWW.HPLCworks.net
858.663 751
Arte et Marte
What was the resolution to this issue? I seem to be having an identical situation. Wavy baseline (~17minute cycles) even after removing column, flushing system, no flow or high flow etc. I was also performing ion chromatography when it started. After backflushing the flow cell, I had a steady baseline for four injections and now I cannot get it back. Pressure is stable. Could be the back flushing happened to rinse different part or the flow cell is now too contaminated? I've run this assay many times before without issue. Please let me know if you found a cause.
Thanks!
23 posts Page 2 of 2

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry