When troubleshooting a system, do remove the column, that makes it easier to find out what hardware if any is not functioning properly.
With column OFF, place instaed a long piece of very narrrow ID tubing to create some back pressure. Then run your baseline. If you do NOT get a good baseline, then your problem probably is , A: an inlet check valve failing, B: solvents need degassing, or [ and I have seen this], a broken plunger.
THIS is 905 of most failing baseline stability ssiues.
Dear Forum:
Please help me with the baseline problem. I set up Waters 2695 Alliance with a 2998 PDA detector using the following parameter:
Column: Whatman Partsil 10-SCX
Column T: 25 celcius degree
Wavelength: 210 nm
Mobile phase: 25% 25 mM KH2PO4 adjusted to ph=4.5; 75% acentonitrile
After I flushed the system (with column and detctor flow cell) with water and 100% methanol, I still couldn't get a stable baseline.
Here is what the baseline looks like:
https://docs.google.com/file/d/0Bw8z-jp ... sp=sharing
Please help trouble shooting the baseline.
Thanks so much!