Was the last trace you posted run after it sat over night with the oven at near ambient temperature (40 °C or so) or did you elevate the oven temperature to "bake out" the system, then cool it before running what you posted?
Only the very old Tekmar systems had a "direct connect" to the GC column. It was a manual system (one sample at a time) and it had a liquid nitrogen trap at the base of the apparatus that sat above the injector. The column was fed all the way up through the GC injector and then through the liquid N2 trap where it was connected to the transfer line via a compression-fitting union. Before desorbing the trap, you hit the liquid-N2 trap with coolant and then inject. Your volatiles were frozen in the trap and at the prescribed time, a ballistic heater was applied to the liquid-N2 trap to liberate the trapped analytes from the column. It worked pretty well. However, the new ones are nicer.