So far we can only eliminate some things:
- at 23% organic you definitely won't have a problem with phase dewetting
- a pH of 2.5 - 2.8 should be no problem for a C18 column concerning bonded phase stripping. If it's indeed a XBridge, I'm sure that it's no problem. I've used an XBridge C18 at pH 1.5, 45°C, thousands of injections with no retention loss.
- the fact that fresh mobile phase doesn't change anything rules out the mobile phase changing over time.
Did you try to wash the column with high organic (or even 100% methanol) to see if the initial retention time can be restored? Be sure to thourougly equillibrate your column with that ion-pairing mobile phase after the washing.
Could it be that actually your samples are fouling the column? There is a way of verifying if the mobile phase or the samples are changing the column. Basically, you keep the mobile phase flowing and inject samples at erratic time points. If you see a correlation between the changing parameter (in your case retention time) and number of injections, the samples are the culprit. If, on the other hand, there's a correlation between analyte retention time and total time (i.e. volume of mobile phase flown through the column) then it's a mobile phase incompatibility.