Yes it uses a 6mm draw out plate and a 1mm inlet liner as well as a 20MX0.18 ultra inert column.  The best part is the BFB_autotune.
I have the larger drawout plate and a 1mm liner, have even used 0.75mm liners before.  I currently use the 40mx0.18 Rtx-VMS column.
I can get enough sensitivity to run regular 524 samples at 100:1 but those UCMR samples have to be run at 30:1 split, even EST has that split ratio in their applications in order to hit the 30ppt MRL needed.
My original problem though doesn't seem to be water related since it is affecting the last few compounds to elute from the column.  Also excessive water should not cause the massive carryover that I see with the original problem.  It seems most likely to be something adsorbing the target compounds within the purge and trap itself.  When clean tubing is used to replace what is present the problem goes away for a couple long sequences then returns shortly thereafter, no changes are done to the GCMS, only the purge and trap.  Something is contaminating the system I just can not figure out what it is.  
I put an indicating hydrocarbon trap on one of the instruments but it has not shown any indication of hydrocarbons entering through the helium supply, so I may have ruled that out.  That would leave reagent water or purge and trap methanol that contains some kind of contaminant that will build up quickly on the sample path tubing.  Reagent water is used to make the standards and blanks and to rinse the sparge tube between samples, while the methanol is used for internal standards and standard spikes.  I wouldn't think the few microliters of methanol would contain enough of a high boiling contaminant to cause the problem so quickly.
The other thing could be sample preservatives.  Ascorbic acid plus HCL or Sodium thiosulfate for the drinking water samples, along with a mix of HCL, or Ascorbic/HCL or Sodium thiosulfate for the waste water samples.  The HCL would be the only thing common to both instruments, but that has been used for decades as a preservative for volatiles samples so the HCL itself I would not suspect, unless it has some contaminate in it I have not identified, but each sample only gets two drops of a 1:1 HCL/DI mixture.
The other culprit could be simply bad samples, probably ran on the soil purge pathway which would not have the foam sensor, but after changing out the sample pathway tubing that should go away.  
Everything that makes sense can be pretty much ruled out, that is what is driving me nuts  