E2M2 is what I am using, they work great still for such old pumps.We ended up switching out the Pfeiffer Duo 2.5 and installing an Edwards E2M2 and inspected the turbo pump (which had a loose bolt). The pressure is now at 3.5x10-5 with a 1.2mL/min flow, the filaments are lighting up and we can get a proper tune. H20<2% and N2 < 0.5%, however I'm still getting over 600 peaks in a scan between m/z 10-700. Am I pretty much looking at an HED failure now or could something else be causing this?
600 peaks but what is your Threshold set to?
You can clean the HED, I have had to do it myself before. Just remove the two screws holding the ceramic part in and pull the connector wire out, should just slide out. Wipe down the metal button with a kemwipe and some IPA. If it is not working at all you should be getting less signal not more I would think since it would not be making the ions take the 90* turn into the multiplier.