New samples run.. Still strange results
Posted: Tue Jul 26, 2005 9:23 am
Now I have tried the following.
I have run 3 standards (8 ppm riboflavin in 20 mM acetate buffer)
The first one was frozen in a vail and then placed in the autosampler
The second was frozen in an eppendorf tube then transfered to a vail that was placed in the autosampler
The third was transfer directly to a vail and placed in the autosampler
Each standard was run 3 times.
What i saw was a HUGE peak (2800 LU )for the first run on the standard frozen in a vail and on the next 2 runs on that vail the peak got smaller(1100 LU and 670 LU) but still to big.
The other 2 standards had peaks that were of the same size (605 +/-2 LU)on all 3 runs of each standard...
Something strange happens in the vail when a sample/standard is frozen in the vail and then run directly from the same vail.
Now i'm going to try to freeze a standard in a vail and when I'm going to run it change the alu cap on the vail to see it the peaks are caused by internal pressure in the vail. I can't see that the pressure should be able to influece the peak size this much but I'm running out of ideas!
JM: I see no other peaks on my chromatogram so I don't know. I don't think I have an injection problem since I can run other samples very reliably.. Just as long as they haven't been frozen in a vail...
I have run 3 standards (8 ppm riboflavin in 20 mM acetate buffer)
The first one was frozen in a vail and then placed in the autosampler
The second was frozen in an eppendorf tube then transfered to a vail that was placed in the autosampler
The third was transfer directly to a vail and placed in the autosampler
Each standard was run 3 times.
What i saw was a HUGE peak (2800 LU )for the first run on the standard frozen in a vail and on the next 2 runs on that vail the peak got smaller(1100 LU and 670 LU) but still to big.
The other 2 standards had peaks that were of the same size (605 +/-2 LU)on all 3 runs of each standard...
Something strange happens in the vail when a sample/standard is frozen in the vail and then run directly from the same vail.
Now i'm going to try to freeze a standard in a vail and when I'm going to run it change the alu cap on the vail to see it the peaks are caused by internal pressure in the vail. I can't see that the pressure should be able to influece the peak size this much but I'm running out of ideas!
JM: I see no other peaks on my chromatogram so I don't know. I don't think I have an injection problem since I can run other samples very reliably.. Just as long as they haven't been frozen in a vail...