Now I have tried the following.

I have run 3 standards (8 ppm riboflavin in 20 mM acetate buffer)

The first one was frozen in a vail and then placed in the autosampler

The second was frozen in an eppendorf tube then transfered to a vail that was placed in the autosampler

The third was transfer directly to a vail and placed in the autosampler

Each standard was run 3 times.

What i saw was a HUGE peak (2800 LU )for the first run on the standard frozen in a vail and on the next 2 runs on that vail the peak got smaller(1100 LU and 670 LU) but still to big.

The other 2 standards had peaks that were of the same size (605 +/-2 LU)on all 3 runs of each standard...

Something strange happens in the vail when a sample/standard is frozen in the vail and then run directly from the same vail. :?

Now i'm going to try to freeze a standard in a vail and when I'm going to run it change the alu cap on the vail to see it the peaks are caused by internal pressure in the vail. I can't see that the pressure should be able to influece the peak size this much but I'm running out of ideas! :(

JM: I see no other peaks on my chromatogram so I don't know. I don't think I have an injection problem since I can run other samples very reliably.. Just as long as they haven't been frozen in a vail... :?
Kind regards
Leadazide

I am very curious for the results; must be your injector or HW Mueller is right and you have problems with precipitation/complex formation or something like that. Good luck and let us know!

Regards Bert

Maybe you are really using a vail instead of a vial....that would explain things. Just a poke.
Seriously: Maybe your vial which was in the refridg had a precipitate which your autosampler kanulae picked up and then injected it? In the others you inadvertendly removed the precipitates, wholly or partially, before injection? Try a very soluble INST.
Having worked on the problem now for some time now I have finally found the reason for the large peaks...

By upscaling the amounts of standard I froze I noticed that the riboflavin was up concentated in certain areals in the sample. And in the vials there isn't enough space for sample to be proberbly mixed using a vortex mixer.

The problem was that I was working with so small amounts of sample and standards that we werbn't able to see this up concentration of riboflavin..

Such a simple thing... :oops:

Thank you everybody for your help and support!
Kind regards
Leadazide
19 posts Page 2 of 2

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry