Folks seem to be forgetting the injection liner volume. I would bet under the conditions stated that the expansion volume of the solvent exceeds the volume of the liner. As I recall a 4mm dia liner volume is about 900ul .
You ask for trouble if you inject 3 ul and split 1:100.

Besidet teh huge risk of backflash and contamination of the injection port gas lines such huge amounts never can give a good split injection because your injection port will not be able to deliver enough energy in the short time allocated for evaporation in a splitted mode. Inject 3x less and split 1 : 33.

Also: use a Precision liner with Quartzwool. This is very important for good evaporation.
You wil get RSD < 1%, guarabteed.
Injection temperature around 250C;

jaap de zeeuw, Restek corporation
17 posts Page 2 of 2

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry