You are saying that the reversed gradient is not good. Please explain as I must say I do not really know why.
Should I better start at a 4-5% and then go for a plateau at 7.5 and then for the wash at 40 or more?
Thanks
How did you arrive at that gradient... was it in the literature or did you come up with it from scratch? What are you particularly trying to accomplish with that drop?
The idea is, at each solvent composition, each analyte will have a certain rate of migration. For each analyte there are solvent compositions that basically don't move it along the column at all, some that will push it along the column at the same rate as the mobile phase, and a narrower range in the middle where its rate changes rapidly.
Normally in RPLC, all analytes move faster at higher organic composition. So you'd expect that once the column has started at 10% ACN, it will elute everything that elutes at any composition lower than 10%. When you drop down to 7.5%, you are essentially slowing down the elution of anything that has not already come off the column.