Re: Back Plunger Error
Posted: Mon Feb 25, 2013 8:33 pm
Thank you for the recommendation - I will take a look at including that chemical as an extra wash.Sticky plunger errors are common on the Agilent autosamplers (personal experience). I recommend adding a second syringe wash solution referenced in the June 2007 issue of LCGC North America, volume 25, No. 6, p. 572. The wash solution is N-Methyl-2-pyrrolidone. Please read the MSDS before use. Since I started using it, my sticky syringe errors have decreased dramatically. If you do use this solvent, you will have to test that it does not interfere with your sample analysis since the peak elutes at about 1.5 minutes on a 30-meter DB-23 column. Hope this helps!