LC/MS/MS Quadrapole Sensitivity Issues
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
17 posts
Page 2 of 2
Thank you for finishing out the thread, oftentimes these issues are left unresolved in this forum. Were you given any advice on preventing contamination of the quads? Like, electrospray needle position, or increased gas flows? On my 4000 QTrap, the service engineer tells me to run at the highest curtain gas flow rate I can without diminishing my sensitivity, and to run the needle at the highest position away from the orifice I can without sacrificing sensitivity, to keep the system as clean as possible.
Time flies like an arrow. Fruit flies like a banana.
-
- Posts: 53
- Joined: Thu Dec 13, 2012 6:29 pm
No problem, I like to be able to finish what I started - because chances are there is someone out there with a similar problem!
Your service rep's advice is basically the same as what mine told me - try to reduce sample load into the quads. With our system we have a diverter valve which I have been trying to program within the method to switch flow from detector to waste (and vice versa) when I know nothing is coming off column. Every little bit that I can do to reduce load on detector is what I was told to strive for.
Overall this is frustrating, these instrument manufacturers design and build these state of the art SUPER sensitive instruments BUT... they can't handle a large sample load (matrix and/or analytes), so we analysts have to dilute and dilute and dilute our samples in order to minimize instrument downtime for maintenance. As it stands right now, I am going to have to vent the instrument monthly to clean the Ion Block of my Xevo TQ-S MS/MS, that procedure takes an entire day to complete and bring system back to normal.
Although i have to admit, perhaps we are just pushing these instruments to the extreme limits of their normal range by taking shortcuts in sample prep, high sample throughput, reducing analysis time, etc...
Your service rep's advice is basically the same as what mine told me - try to reduce sample load into the quads. With our system we have a diverter valve which I have been trying to program within the method to switch flow from detector to waste (and vice versa) when I know nothing is coming off column. Every little bit that I can do to reduce load on detector is what I was told to strive for.
Overall this is frustrating, these instrument manufacturers design and build these state of the art SUPER sensitive instruments BUT... they can't handle a large sample load (matrix and/or analytes), so we analysts have to dilute and dilute and dilute our samples in order to minimize instrument downtime for maintenance. As it stands right now, I am going to have to vent the instrument monthly to clean the Ion Block of my Xevo TQ-S MS/MS, that procedure takes an entire day to complete and bring system back to normal.
Although i have to admit, perhaps we are just pushing these instruments to the extreme limits of their normal range by taking shortcuts in sample prep, high sample throughput, reducing analysis time, etc...
BHolmes
Any problem worthy of attack, proves its worth by hitting back...never give up!
Any problem worthy of attack, proves its worth by hitting back...never give up!
17 posts
Page 2 of 2
Who is online
In total there are 139 users online :: 0 registered, 0 hidden and 139 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 139 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 139 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science