Page 2 of 2
Re: GC-Negative Peak
Posted: Mon Feb 11, 2013 4:26 pm
by chromatography1
Yes, my column comp is on also I set up zero on the GC.
Thanks.
Posted: Mon Feb 11, 2013 4:54 pm
by GasMan
If you do not start a run on the 5890 by pressing 'START', but just increase the oven temperature to the final temperature of your temperature program, does the baseline still go negative?
Gasman
Posted: Mon Feb 11, 2013 5:07 pm
by chromatography1
The base line changed with temperature if I pressed, “START “without injection.
I also tried isothermal test, I pressed “START” without injection and without temperature change, the base line moved to negative side.
Posted: Mon Feb 11, 2013 7:23 pm
by GasMan
You need to know what happens if you do NOT press Start run, but just put the oven temperature up to the final temperature of your temperature program. This will tell us if it is a 'chromatography' problem if you still see the baseline go negative.
I have the feeling that you have a baseline compensation run stored in your GC, and the signal chosen is something like 'A -ColComp1'. If the compensation run was made on a column that had not been conditioned correctly, then it is possible that the compensation signal is greater than the present bleed of your column. This will give you your negative output.
Do what I say at the top of this message, this will tell you immediately where the problem is. If there is no negative signal by just increasing the oven temperature, check what your signal entry is. If it has the '-ColComp' as part of the signal, remove this by just entering A or B for your signal.
Gasman
Posted: Mon Feb 11, 2013 8:56 pm
by chromatography1
Thanks for all helps,
After removed “Col Comp” the signal has no baseline problem.
Posted: Wed Feb 13, 2013 7:20 pm
by Yama001
Nice catch on the column comp setting folks.