Special thanks for you Tom for the Idiom and for Sherlock Holmes quote!
----
I used a brand new column today for five injections of ACN using the same method described in the first post, with no guard column. I got the below chromatogram for the last four injections. (I excluded the first one as it was not stable enough and I used it as a preparation for the new column).

The chromatograms here are almost identical to what I got after cleaning the previous column (point 8th in my first post).
I also noticed the same peaks in both set of chromatograms for the two columns. I checked the peak areas of the largest TWO peaks at 25.6 min and 26.0 min in today's set of chromatograms and got the following:
RUN #-PEAK AREA FOR 25.6 min-PEAK AREA FOR 26.0 min
#1-1967-945
#2-2296-1148
#3-2609-1274
#4-2761-1323
There is a clear increase in the peak size from an injection to the next. I am feeling that there is a build-up of a UV-absorbing material from somewhere. What do you think?
It might be possible that the problem is a combination between a detector problem and a contamination.
Before installing the new column I thought it might be better to clean the injector with some large injections, so I injected 10 replicates of 100uL injections of 100% ACN (with no column in place). The run was isocratic at 50% ACN/water, 0.5 ml/min. I got a high peak (> 0.002 V) in less than 1min. Just to have more fun, I injected 50uL and 10uL to see what I may get. The results are shown below.

I was wondering why I got a peak for only ACN! I thought the reason might be the differences in the characteristics between the injected 100%ACN and the mobile phase (50%ACN), so I injected 100uL of 50%ACN and 100%Water and got the below:

I wonder if this is a NORMAL effect, and also why to get a peak for the mobile phase injections?
After disconnecting the new column I ran an injection of ACN using the same gradient method in post 1 with NO COLUMN in place (tubing instead). I got a clear chromatogram with a high peak in front of the gradient as the chromatogram in my THIRD post.
Is there anything to worry about here?
Sorry for the LONG story but I am getting lost!
Thanks.