Re: Y-Presstight connectors

Posted: Thu Nov 29, 2012 1:02 pm
by Bigbear
I re use the MXT connectors, I just sonicate them in solvent for a while.

Re: Y-Presstight connectors

Posted: Thu Nov 29, 2012 2:22 pm
by chromatographer1
You are wise as well.

I was glad to see that gserrano_84 confirmed my experience with the press-tights with a GC-MS detector. The polyimide is not impermeable to air even if no gaps in the seal are achieved.

This small contamination of air is not a deal breaker but your column will last longer without the trace contamination of air. For this and other reasons when connecting my Headspace Columns to the transfer line I always used the Restek deactivated unions with a steel body and with graphite ferrules.

best wishes,

Rod