Try a lower incubation temperature and a shorter equilibrium time - you are generating high pressures in the vials by equilibrating at 105C, and 45 min gives a long time for small leaks to affect your analysis.
However, this still does not explain why the first vial is consistently higher than the others. Try running a blank vial (everything except the benzene) as the first vial in the series to eliminate the possibility of carryover.
Have you tried the experiment of filling and capping each vial just before it starts to equilibriate ?
You may be tightening the caps too much - does the top of the septum have a dent in it after you have crimped the cap ? If it does you have probably split the teflon liner, and the sample is exposed to the silicone, which is absorptive.
Peter
‘Try a lower incubation temperature and a shorter equilibrium time’, I agree with u for good rsd. But peak area may also decrease, maybe a influence to sensitivity。Mean smaller benzene peak,and carbon tetrachloride in class1 standard smaller peak,hard to pass the signal-noise-ratio requirement。That’s all my imagine,actual situation need to try to know。
The vial before the six vial is class 2A standard,no benzene in it。I see its chromatogram,corresponding position very small peak,area 0.7(before 2A is class1 standard,containing benzene). So small that it doesn’t matter to my assay,can’t be the reason to harm rsd。
I thought 0.1ppm benzene water solution not steady,if not parallelly transfer to vial and cap,will bring more difference,maybe bigger rsd。So never try as you say。And the manner will tire person,making person in pressure,time between injections about 66min。must late for regular bus,have to extra work。
Occationally tightening the caps too much,I really found some vial septa side having dent。
What like is ideal state to cap。I thought regular dent all around the aluminum cap is perfect,this mean septa is compress,thickness lessen,tighter。Or absolutely flat at the side of aluminum cap,is best?there should be some moulage at the side of aluminum cap(I mean the position, when turn cap, fingers hold position), for ideal operation?instinct I doesn’t think it’s decisive factor to terrible rsd.
A new idea strike me. Maybe the benzene water solution of 0.1ppm, not homogenous。Benzene density lower than water,maybe part dissolved,part in other style floating in upper bottle。When draw solution,different depth and time,contribute to bad rsd。Maybe the solution prepared keep still 1 hour before transferring to vial,helpful,although maybe loss some benzene not dissolved。
Thanks for ur attention。
Best wishes。