PROBLEM WITH Gas Chromatograph 7890A
Discussions about GC and other "gas phase" separation techniques.
25 posts
Page 2 of 2
I think a blockage in either FID jet would cause longer retention times in both columns. If you have spare FID jets then I would try replacing these, it is easy and quick. If that doesn't work can you post chromatograms from both FIDs, before and after the problem started?
Where can I buy the kit they use in CSI?
-
- Posts: 11
- Joined: Tue Sep 04, 2012 8:00 am
Thank you for your suggestion...I think a blockage in either FID jet would cause longer retention times in both columns. If you have spare FID jets then I would try replacing these, it is easy and quick. If that doesn't work can you post chromatograms from both FIDs, before and after the problem started?
I can't replace the jets because I haven't spare but I will clean them. This week, I will try clean them, when I will do I will comment that has occurred.
I will write as soon as possible!! Thank you for your help...
-
- Posts: 377
- Joined: Wed Aug 24, 2011 2:10 pm
They can be difficult to clean, so sonicate in water then methanol, and try to poke out the orifice with a small wire, I know a guy who uses a strand of a guitar string. You really need spares of these jets, if you swap one good one for another then you can just keep it as another spare, but the down-time of waiting for an order is too long, and clenaing doesn't always work.
Where can I buy the kit they use in CSI?
-
- Posts: 41
- Joined: Wed Jun 17, 2009 6:02 pm
When the FID peaks are becoming even more delayed over time, what is the peak shape like? Are the peaks becoming wider?
-
- Posts: 18
- Joined: Thu Jan 26, 2012 4:03 pm
maybe you need trim the column which is in the detector side.
-
- Posts: 11
- Joined: Tue Sep 04, 2012 8:00 am
Thank you very much for your help
Yes, the peaks were becoming wider and fatter. The solution has been to change the column.
Yes, the peaks were becoming wider and fatter. The solution has been to change the column.
-
- Posts: 11
- Joined: Tue Sep 04, 2012 8:00 am
Hello!
In my laboratory, we have two gas chromatograph 7890A. Now, the problem I have in the other chromatograph 7890 A (not in the chromatograph where I observed the movement of chromatographic peaks).
Now, the problem is that in the middle of a sequence, the FID turn off. I think that the problem aren't the air and the H2 because if the problem were these gases two FID shut down.
In this gas chromatograph, we have two FID detectors.
Please, help meeeeeeee!! Thank you very much!!!!
In my laboratory, we have two gas chromatograph 7890A. Now, the problem I have in the other chromatograph 7890 A (not in the chromatograph where I observed the movement of chromatographic peaks).
Now, the problem is that in the middle of a sequence, the FID turn off. I think that the problem aren't the air and the H2 because if the problem were these gases two FID shut down.
In this gas chromatograph, we have two FID detectors.
Please, help meeeeeeee!! Thank you very much!!!!
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
A clean detector that has good gases set at the proper flows, having no defective parts, bent or blocked, will not fail to maintain a flame.
Would this be the GC that had the bad column? (which may have bled deposits into the FID?)
Start cleaning and checking your parts.
best wishes,
Rod
Would this be the GC that had the bad column? (which may have bled deposits into the FID?)
Start cleaning and checking your parts.
best wishes,
Rod
-
- Posts: 11
- Joined: Tue Sep 04, 2012 8:00 am
A clean detector that has good gases set at the proper flows, having no defective parts, bent or blocked, will not fail to maintain a flame.
Would this be the GC that had the bad column? (which may have bled deposits into the FID?)
Start cleaning and checking your parts.
best wishes,
Rod
This GC isn't that had the bad column. It´s another GC 7890 A, but I think that first I would have to clean the FID jet???
Thank you very much for your answer!!
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
A good place to start.
Rod

Rod
25 posts
Page 2 of 2
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science