I think we're using terms in different manners. For me, "system peaks" (some call them "ghost peaks")refer to anything that's visible during a blank run (blank meaning no injection - some are referring to solvent injections as blanks). In isocratic elutions, you usually don't have any system peak. With gradient elution, they're quite common due to traces of eluent impurities that accumulated on the column at the beginning and elute when the eluent gets strong enough.
Any peak visible with a solvent injection that's not a system peak as described above I'd call a solvent peak. This is mainly at t0 ("Solvent front", "t0 noise"), but may be also visible later in the chromatogramm.
The thing you described ("System peaks arise due to the presence of analyte in the sample, competing with the species present in the eluent") sounds what I would call "vacancy peaks". They might be positive or negative depending on the absorption characteristics of the analyte and the eluent and the detection wavelength. I think this is what you were referring to?