I vent, clipped and rehooked the column to the MS and tried it again. No Change.Due to the way that back-pressure regulated inlets (which I am pretty sure the Agilent has) are plumbed, there should be flow out of the split vent all the time. Splitless transfer to the column is generated by rerouting the gas flows inside the plumbing - this makes it infuriatinlgy difficult to troubleshoot back-pressure regulated inlets.I thought it might be the split valve stuck open so stuck a flowmeter on the split vent and it read 0 ml/min for 1 min while doing a splitless injection so no help there.
Try a split injection - about a 50:1 split of a 1 mg/ml solution will give you 20 ng per peak which should be easily detectable.
Peter
I tried injecting a 4 ug/ml test standard 50:1 split and didn't really see anything (with a split liner). I also injected the same mixture 1ul splitless 1 min before purge and got pathetic whimpy little peaks I should see peak heights ~100,000 instead I am seeing barely 1000.
tune and air check still look fine. I have flow through the column (I dipped the end into a vial of methanol and saw a fast stream of small bubbles. I am running out of theories. THe inlet is clean and all the consumables are new (gold seal, septa, liner)