Re: Looking for Waters 996 DAD Service Manual or Error Codes

Posted: Mon Jul 30, 2012 6:47 pm
by sig
Today I have tested my detector by Millenium diagnostic tests. Surprisingly, it passed all test except two - lamp was ignited at the second run ( 400 h on counter) and 485 nm test was failed ( delta 2.5 nm). I hope I can work with it for the beginning, and then replace the lamp.

Re: Looking for Waters 996 DAD Service Manual or Error Codes

Posted: Mon Jul 30, 2012 6:50 pm
by remesquaddie
These 996 detectors almost always fire up the lamp on the 2nd attempt. , even on new lamps. that is usually not a problem.

Re: Looking for Waters 996 DAD Service Manual or Error Codes

Posted: Mon Jul 30, 2012 6:51 pm
by remesquaddie
You might want to run a caffeine standard to verify the wavelength accuracy. Maxima should be at 273.3nm

Re: Looking for Waters 996 DAD Service Manual or Error Codes

Posted: Mon Jul 30, 2012 7:15 pm
by sig
Thank you ! I will finish with pumps and software (Gilson 305/306, 232 autoinjector, UniPoint software, 506C interface) and will make all standard tests - detector linearity response, wavelength accuracy, column performance tests. I worked much with Gilson-Rainin-Dynamax, but this is my first experience with Waters. Unfortunately, old Gilson's DAD works bad, but I need such a detector due to my specific tasks.

Re: Looking for Waters 996 DAD Service Manual

Posted: Wed Jun 14, 2017 12:00 pm
by wschrabi2
HI Experts,
I am looking for the Wayter 996 PDA DIP Switches Address Definition for the IEEE-488 address. In the manual there is written: ".. Refer to Millennium Chromatography Manager System Config Guide Table 2-2, for the correct IEEE-488 DIP switch settings for each address." Unfortunately this Guide is not handy here and I would like so use the NI GBIP-USB-HS Adapter to use it on an Ezchrome HPLC System. At this Manual http://www.ecs.umass.edu/eve/facilities ... manual.pdf and here https://www.google.at/url?sa=t&rct=j&q= ... TUgU8RHkzQon page 18 there is: "... Section 2.3.1 Connecting IEEE-488 Devices, ...", but even this Manual is not handy. I have found in page2-14 this info in http://www.waters.com/webassets/cms/sup ... 9707rb.pdfMoreover I would like to know, which Firmware version our PDA has, so that we can check if the required version 1.8 meets the requirements. ANy HELP sofar? Thanks in advanced. Walter

Re: Looking for Waters 996 DAD Service Manual or Error Codes

Posted: Thu Jun 29, 2017 6:20 am
by koiseptember
Thank you very much for this information. I will try to connect with Waters technical support in my country, but I strongly suspect that they will recommended me to buy new Breeze or Aliance (approx. $150000-200 000 in Russia)