If you leave the vial number blank (no entry) the Agilent 1100 will perform a no injection run. The valve will NOT move from inject to load to inject as in a normal run which eliminates the injector rotor/stator as the source of the problem. The loop is always inline unless you run an injector program.
A. Carl Sanchez
check this
http://chromforum.org/viewtopic.php?t=7570

Hope it gives u some idea
we had a weird case of a negative peak in the chromatogram kind of like that.
it was stupid really, and we have no idea why it was the behavior we had,
but someone made the runs on a waters RI while the purge was "on" and it gave a negative peak somewhere
I would say, that you have contamination in sampler area (Loop, metering device, needle, needle seat or valve). Try heating water to around 70 Degree celsius, bypass the degasser, and flush the system through the valve position in main pass. Use zero-dead volume insted of column and set the thermostat to max.
19 posts Page 2 of 2

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry