Re: Problems with Kinetex C18, 1.7 µm
Posted: Sat Apr 21, 2012 7:08 am
The QC department where the method has problems (that could be related to column batch variations, we don't know), also has a Waters Acquity instrument. I have changed the capillaries to and from the column, but I have not done more than that on their instrument.Mattias,
I'm sure you are aware of this but even a UHPLC system must be optimized in order to realize the improvement in performance provided by the highest effieciency columns available today. The QC results for the columns in question provide a good estimate of the performance under realistic conditions. These results are generated on standard UHPLC instruments (>100 instruments) so users should not have a problem reproducing these results when using similar equipment. If you are unable to reproduce the QC results then the system must be optimized. Unfortunately, many UHPLC system are not capable of showing the true performance of these ultra-high effieciency sorbents without careful optimization.
The trooble shooting point towards the columns, but I feel that it is very strange that I every column that I buy and start to use works fine (for at least 5-6 batches of columns). QC have now two batches of columns with the same problem (I guess I should ask them to send one of their new columns to me).
Back to the instrument: are there other things that need optimisation, or that could make peaks splitted? Their Acquity is almost new, and my instrument is from 2006. It appears as I generally get higher plate counts than QC.