Cleaning MSD

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

17 posts Page 2 of 2
The methylene chloride helps to ensure that any less polar compounds are stripped from the source parts. You can try abreviated cleaning procedures, but you need to be sure that this abreviated cleaning works for all of the sample types run on the instrument. If the instrument is a "general use" type of instrument, you have no idea what the next sample will be or what the other users' samples left in the ion source. An abreviated cleaning procedure may be more appropriate on an instrument that runs a particular type of sample.
Blow off parts with canned air if desired then air dry for 10 minutes. Done.
I'd be inclined to skip the canned air - propellants vary and you might end up contaminating your source. You could use zero air if you want to blow dry your source but a quick bake in the oven seems just as easy to me.
17 posts Page 2 of 2

Who is online

In total there are 54 users online :: 2 registered, 0 hidden and 52 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot], Semrush [Bot] and 52 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry