lmh » Thu Nov 15, 2012 11:23 am
Are you sure it's 278 and not nearly 279? 279 would have been a common phthalate contaminant of solvents, a plasticizer. The ion trap isn't very accurate, so it might come out at 278.9
If you've got a surveyor hplc system, the injection mechanism can be a bit difficult. It's almost impossible to remove all bubbles from it, and it tolerates a few bubbles, but if it runs out of needle wash it stops injecting altogether. If the needle ever gets partially blocked, or the drive mechanism on the syringe gets dirty, the drive motor will slip a bit, and it loses coordination of how far up or down it is. This matters because it uses a double-plunger concentric syringe, and how far up or down it is determines which plunger is moving. If it thinks it's pipetting with the middle bit but is actually using the larger bit, it will think it's pushing the sample neatly into the sample loop, but actually it's pushing it straight out the other side and into the waste!
According to wikipedia, Alachlor has a neutral mass of 269... it's not got huge numbers of easily ionised groups, so it might not give a huge signal? If I were you, I'd go back to syringe injection, but if you can't do the teeing thing because your syringe is too leaky, just inject directly into the source at 5uL/min or so, and see if you can detect it at all. You'll need to reduce the sheath gas substantially at this low flow rate, and no aux gas, but otherwise the tune parameters can stay the same.