From the beginings I thought TIC = total ion current http://tinyurl.com/bmm7fkpI always thought that "TIC" stands for Total Ion Chromatogram. It is a basic principal of measurement that a number without a unit is not a measurement. The SI unit for counting entities such as ions is the Mole.
Peter
Hi Peter. Can't be Mole. How can you imagine 2 000 000 or 3 000 000 moles of any ion in MS chamber ?Hi dblux
If TIC = Total Ion Current then the unit is Amperes.
The MS software that I have (Bruker) shows signal strength in "counts" not relative abundance (which prompts the question; relative to what ?).
Since in ordinary GC-MS all the ions are singly charged, the Coulomb would be an alternative to the Mole in the SI.
Peter
Yes it is, but abundance of what ?, and what are the units ?. The heading has Total Ion Chromatogram, not Total Ion Current by the wayHi Peter. Can't be Mole. How can you imagine 2 000 000 or 3 000 000 moles of any ion in MS chamber ?Hi dblux
If TIC = Total Ion Current then the unit is Amperes.
The MS software that I have (Bruker) shows signal strength in "counts" not relative abundance (which prompts the question; relative to what ?).
Since in ordinary GC-MS all the ions are singly charged, the Coulomb would be an alternative to the Mole in the SI.
PeterI can't imagine that, but what I can omagine does not alter the fact that the SI unit for defined entities such as the ion fragments in an MS is the Mole, with the entities to be specified. If the entities are not specified, since we have a charged particle the Coulomb might be appropriate. That the vertical scale on the chromatogram of MS software has huge numbers on it simply means that the units on the screen are not SI units
ChemStation shows signal intensity in "counts" as well. And counts is related to number of ions hiting MSD detector (horn). However that number is multiplied by electron multiplier.and then converted and digitized before any of us get to see it. In terms of quantity of analyte present, the signal intensities are arbitrary until they have been converted to masses or Moles via a calibration of signal (usually in terms of peak area) vs known analyte quantity
PS - relative to constant (factor or how it is called in English) of the detector a drawback of MS (compared to e.g. an FID) is that the response per unit analyte is not constant
PS 2 - if FDA is reliable source of information see 1st chromatogram in an article:
http://www.fda.gov/Food/ScienceResearch ... 113209.htm
Y-axis is abundance !
Abundance of ions; counts is the measure of abundance and so the unit is ion, but this is not an SI unit, which is what Jaap was asking for, it is dimensionless (or cps=Hz)the c in cps is cycles per second, not counts really doesn't matter.
Above statement is only according to my humble opinion. If there is a reliable source of knowledge published as eg. manual or help file to MS CDS which may indicate I'm wrong I would change my opinion.
Yes, Total Ion Chromatogram. It's a short form of Total Ion Current Chromatogram. It's only naming convention.
PS - you are encouraging me to read the book from my 1st link, well maybe in the future
Peter, ion is not an unit !, it's a particle bearing electric charge. Count is unit and it has no SI dimension like number of digits at your hand. That's like counters measure numbers of charged particles.and so the unit is ion, but this is not an SI unit, which is what Jaap was asking for
Noooo, the c in cps is not cycles per second, it's simply cycles, combined with "ps" gives Hz. But dimension of abundance expressed as counts really doesn't matter and no one needs it neither for quality nor quantity analyses.,...(or cps=Hz)the c in cps is cycles per second, not counts...
These people are much better at explaining the SI than I am http://www.bipm.org/ - the plans to reform the whole system, and discussions on what the units should be for counting are particularly relevant to this discussion.Peter, ion is not an unit !, it's a particle bearing electric charge. Count is unit and it has no SI dimension like number of digits at your hand. That's like counters measure numbers of charged particles.and so the unit is ion, but this is not an SI unit, which is what Jaap was asking for
Noooo, the c in cps is not cycles per second, it's simply cycles, combined with "ps" gives Hz.you are right, my bad, but what has cycles per second got to do with counting ions ? But dimension of abundance expressed as counts really doesn't matter and no one needs it neither for quality nor quantity analyses.,...(or cps=Hz)the c in cps is cycles per second, not counts...
Hi,
I guess the "counts" is a part of term "Total Ion Counts (TIC)". If yes, there is no unit for "counts" then. It is just a natual number. However, the y axis units, ABUNDANCE, for either TIC, (time vs. abundance) or mas spectra (m/z vs. abundance) uses unit of percentage or ppm (1% = 10,000ppm).