The method i am following is current ISO practice (5508:1995 and 12966-2:2011) with a 1:15 dilution stage at the end
Peter:
Yes, I am still experencing discrimination against the heavier c-chains
My injector settings are
Sample Volume: 1.0ul
Pre-Inject Delay: 0s
Post-Injec Dwell: 1s
Injection Speed: "1" an arbitrary value given by the clarity software that i can only describe as not fast... increasing number = increasing speed, up to a value of 5 i think
There are 5 solvent flushes prep and post injection of 3ul Iso-octane
BigBear / Krickoss:
The solvent expansion in the liner should not exceed that of the liner volume. I have checked this using the calculation on the Restek site (desilanised Restek precision liners used with glass wool insert)
Alexandre:
There is no hump/baseline noise increase with inj temp increase, increase in oven temp obviously does give an increase in baseline but no humps appear which may be masking peaks
Once again thanks for the help, its much appreciated.