Re: Ethanol as ISTD in Spirit Drinks
Posted: Thu Jan 19, 2012 11:03 am
Continuation
Dear Peter,
verification of method stability against dilution of testing samples was been done by the next way. Three reference ethanol-water solutions were analyzed after dilution with water in ratio 1:1 and 1:3. The obtained results are presented below in the Table.
Compound___Known_conc._of_i-th____Measured._c.___Relatively____Measured._c._____Relatively_
__________compound_in_standard__after_dilution_1:1_discrepancy_after_dilution_1:3__discrepancy
_____________(mg/L)_______________(mg/L)___________%__________(mg/L)__________%_____
acetaldehyde____10.11______________10.34___________+2.2__________10.50_________3.8____
_______________99.64______________97.28___________-2.4___________97.40________-7.3___
_______________497.6______________483.3___________-2.9___________473.1________-4.9___
methyl_acetate__10.00______________10.25___________+2.5___________9.78_________-2.2___
_______________100.0______________92.76___________-7.2___________89.17________-10.8__
_______________500.0______________463.7___________-7.3___________452.4________-9.5___
ethyl_acetate____10.47_____________10.18___________-2.8___________10.63________+1.6___
_______________104.7______________100.0___________-4.5___________95.46________-8.8___
_______________523.4______________489.6___________-6.5___________477.3________-8.8___
methanol________103.2_____________97.99___________-5.0___________95.18________-7.8___
_______________1005______________921.9___________-8.3___________904.1________-10.0__
_______________5013______________4654____________-7.2___________4514________-9.9___
2-propanol______11.80______________11.63___________-1.2___________10.56_______-10.4__
_______________103.2______________97.86___________-5.2___________93.13_______-9.7___
_______________509.4______________479.5___________-5.9___________463.7_______-9.0___
1-propanol______10.21______________10.36___________1.5____________10.01_______-1.9___
_______________102.1______________98.00___________-4.0___________96.23_______-5.7___
_______________523.4______________482.9___________-5.8___________483.2_______-7.7___
isobutyl_alcohol__10.00______________10.42___________+4.2__________10.35_______+3.5__
_______________100.0______________96.87___________-3.1___________94.32_______-5.7__
_______________500.0______________480.1___________-4.0___________471.5_______-5.7__
n-butanol_______10.00______________10.17___________+1.7___________9.98________-0.2__
_______________100.0______________97.02___________-3.0___________95.21_______-4.8__
_______________500.0______________482.9___________-3.4___________475.2_______-5.0__
isoamyl_alcohol__10.37______________11.28___________+8.8___________10.35_______-0.2__
_______________103.7______________103.0___________-0.6___________99.52_______-4.0__
_______________518.2______________509.0___________-1.8___________500.6_______-3.4__
Estimated volume of relative accuracy for proposed method followed from analysis of the obtained experimental in accordance with "ISO 5725-1994. Accuracy (trueness and precision) of measurement methods and results" does not exceed 11 %.
Thousands of testing laboratories over the world every day carry out gas chromatographic analysis of volatile compounds in spirit drinks. They may test this approach in their real practice. It is important to note that there is no need to perform any additional measurements. This method could be tested while performing current measurements with existing instrumentation and calculations could be done in parallel according to two different methods: using traditional ES method (GOST R 51698-2001. Vodka and ethanol. Gazochromatography express-method for determination of toxic microadmixtures content, 2001) or according to (Commission Regulation (EC) No 2870/2000 of 19 December 2000 laying down Community reference methods for the analysis of spirits drinks, 2000) with pental-3-ol as ISTD and using ethanol as ISTD.
I promise my prompt help on e-mail: chere@inp.bsu.by
Best Regards,
Siarhei
Dear Peter,
verification of method stability against dilution of testing samples was been done by the next way. Three reference ethanol-water solutions were analyzed after dilution with water in ratio 1:1 and 1:3. The obtained results are presented below in the Table.
Compound___Known_conc._of_i-th____Measured._c.___Relatively____Measured._c._____Relatively_
__________compound_in_standard__after_dilution_1:1_discrepancy_after_dilution_1:3__discrepancy
_____________(mg/L)_______________(mg/L)___________%__________(mg/L)__________%_____
acetaldehyde____10.11______________10.34___________+2.2__________10.50_________3.8____
_______________99.64______________97.28___________-2.4___________97.40________-7.3___
_______________497.6______________483.3___________-2.9___________473.1________-4.9___
methyl_acetate__10.00______________10.25___________+2.5___________9.78_________-2.2___
_______________100.0______________92.76___________-7.2___________89.17________-10.8__
_______________500.0______________463.7___________-7.3___________452.4________-9.5___
ethyl_acetate____10.47_____________10.18___________-2.8___________10.63________+1.6___
_______________104.7______________100.0___________-4.5___________95.46________-8.8___
_______________523.4______________489.6___________-6.5___________477.3________-8.8___
methanol________103.2_____________97.99___________-5.0___________95.18________-7.8___
_______________1005______________921.9___________-8.3___________904.1________-10.0__
_______________5013______________4654____________-7.2___________4514________-9.9___
2-propanol______11.80______________11.63___________-1.2___________10.56_______-10.4__
_______________103.2______________97.86___________-5.2___________93.13_______-9.7___
_______________509.4______________479.5___________-5.9___________463.7_______-9.0___
1-propanol______10.21______________10.36___________1.5____________10.01_______-1.9___
_______________102.1______________98.00___________-4.0___________96.23_______-5.7___
_______________523.4______________482.9___________-5.8___________483.2_______-7.7___
isobutyl_alcohol__10.00______________10.42___________+4.2__________10.35_______+3.5__
_______________100.0______________96.87___________-3.1___________94.32_______-5.7__
_______________500.0______________480.1___________-4.0___________471.5_______-5.7__
n-butanol_______10.00______________10.17___________+1.7___________9.98________-0.2__
_______________100.0______________97.02___________-3.0___________95.21_______-4.8__
_______________500.0______________482.9___________-3.4___________475.2_______-5.0__
isoamyl_alcohol__10.37______________11.28___________+8.8___________10.35_______-0.2__
_______________103.7______________103.0___________-0.6___________99.52_______-4.0__
_______________518.2______________509.0___________-1.8___________500.6_______-3.4__
Estimated volume of relative accuracy for proposed method followed from analysis of the obtained experimental in accordance with "ISO 5725-1994. Accuracy (trueness and precision) of measurement methods and results" does not exceed 11 %.
Thousands of testing laboratories over the world every day carry out gas chromatographic analysis of volatile compounds in spirit drinks. They may test this approach in their real practice. It is important to note that there is no need to perform any additional measurements. This method could be tested while performing current measurements with existing instrumentation and calculations could be done in parallel according to two different methods: using traditional ES method (GOST R 51698-2001. Vodka and ethanol. Gazochromatography express-method for determination of toxic microadmixtures content, 2001) or according to (Commission Regulation (EC) No 2870/2000 of 19 December 2000 laying down Community reference methods for the analysis of spirits drinks, 2000) with pental-3-ol as ISTD and using ethanol as ISTD.
I promise my prompt help on e-mail: chere@inp.bsu.by
Best Regards,
Siarhei