From your configuration, I'm guessing your system max pressure is 600 bar, right? If so, you don't have a UHPLC (typically capable of 1000-1200 bar, optimized to have low system volumes, etc). You have what Agilent used to call a Rapid Resolution HPLC. You can use their RRHT columns (Rapid Resolution High Throughput) with your system, or their Poroshell 120 columns (or similar). This system was Agilent's first attempt at cracking into the UHPLC world dominated by Waters at the time. UHPLC is basically used to describe systems that can realize the efficiencies provided by sub-2-micron particle size columns (and also superficially porous columns like Phenomenex's Kinetex, AMT's HALO, and Agilent's Poroshell columns).
But I digress. Like Tom said (and like I said previously, as well as Alexandre) - you probably aren't going to get to your desired detection limit using UV detection, regardless of the improvements you make. To get as close as possible, you can do the following:
1. Use the most UV-transparent mobile phase you can - something like acetonitrile-water.
2. Use as small a particle-sized column as you can (the 1.8-micron RRHT columns can be used on your system - something like a ZORBAX SB-C18 RRHT might work).
3. Ask Agilent about a low-volume flow cell for the DAD that will help you maximize UV-sensitivity (I'd guess they'll suggest something in the micro to nano range).
Micro: 1.7ul, 6mm path length, 400bar P/N G1315-60024
Semi-Nano: 500nl, 10mm path length, 50bar P/N G1315-68724
Nano: 80nl, 6mm path length, 50bar P/N G1315-68716
4. Make sure you are monitoring the best wavelength for each compound. You can monitor up to eight signals in the DAD - use them!
5. Inject as much sample as you can to maximize your signal-to-noise ratio, without destroying your chromatography.
6. Reduce your extra-column volume as much as possible to reduce peak dispersion.
All that being said, if you get 10 times more sensitivity than you have now, I'd be impressed. But 100? Highly doubtful. For a more in-depth look at optimizing your system, see the following:
http://www.chem.agilent.com/Library/use ... _ebook.pdf