What has worked for our lab is the following (I've only been doing GC work for just over a year now):

Cut retention gap, wipe with acetone dampened kimwipe, apply a small dab of resin (Agilent PN: 500-1200) near the end of the column (a few mm), insert into the union and hold in place for 20-30 seconds.

Repeat with the column end. Once both sides are seated in the glass union, set the inlet and oven temps to method starting conditions, but keep the helium flow/pressure about 40% of it's starting condition (ie. if 100 kPa, set to 40 kPa) is for 30-60 minutes. Then increase the flow to normal operating levels.

We keep the resin at 4C when not being used, and let warm to room temp before applying.

This has provided for reliable seals using glass unions. Another thing of note. We were using Agilent Universal glass unions, PN: 705-0825, and even though the part number is the same, the most recent batch we got was physically different. The current ones have "fluted" ends and if you place a .530mm fused silica retention gap in one end, it seats 9mm inside from one end and only 5mm from the other end, and the earlier ones were "flat" on the ends and the RG seated ~8mm from either end.

Where this distance became an issue is if you do add the resin, you have to make sure it's close enough to the end of the RG to also "stick" to the union by making contact. If it's too far from the end of the RG, it may not come in contact with the glass union and form as good of a seal (if any).

Just saw this post, so my answer may be a bit late. :D
Re: the resin, yes, it needs to close enough to the end of the column to seal, but not so close that it gets in the end of the column and blocks it.
Good day,

I have seen and tried a number of strategies using the polyimide. I was not having any luck using the Siltek unions with the Siltek retention gap, a little too "slippery" of a combination, (that was causing some grief). I have recently started using a de-activated glass union with de-activated retention gap from Agilent without the resin and so far it seems to work quite well. I do not always get the column and/or retention gap to "fuse" to the union however I do get it to seal reasonably well! I have tried the strategy you have suggested, with success however I do have concerns about getting it in the column and prefer to go without the resin.

Thank you very kindly for your suggestions, I may try it again!

Cheers,

Tom
18 posts Page 2 of 2

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry