Definitely an interesting topic. As the definition of robustness testing was put to me, "small but deliberate changes, but not so large to explore points of failure".
I recently finished a gradient validation and below are the parameters we examined.
Initial organic % gradient value
Final organic % gradient value
column temperature
buffer ionic strength
initial wavelength
wavelength change (we're deliberately changing the wavelength during the run because of component wavelength intensity)
flow rate
These were examined using a design of experiment program, so multiple parameters were changed simultaneously. The max and min values were typically 5-10% of the nominal value. Most of these are instrument parameters. Why? Because parts are moving and they wear out. It's possible the proportional valve may not be perfectly accurate and get back to my starting point of 5 /95. There may be some variability in temperature due to the oven itself, maybe the location of the column in the oven or if there is influence of the room temperature. Maybe the pump piston seal is starting to leak so the flow rate may be a bit off, but not enough to shut down the system.
Remember, your suitability parameters should catch something if there is a problem. Robustness testing provides the ability to make notes of caution within your test method if a certain parameter is sensitive and to avoid going beyond that threshold.