I am using a binary agilent 1100 series pump. When purging the lines A and B at 5 mL per min to waste I heard clicking from the pump A (could feel it clicking too). It cause spikes in pressure but went away after a while. I rinsed those lines with a lot of water.

After purging all lines, sample cell, and reference cell with water, ethanol, water again, and after cleaning the column with 10%DMSO and re equilibrating with mobile phase, my noise went from 500 nRIU to 5 nRIU. However, when I purged the reference cell my noise went up to 50 nRIU and now it's up to 100 nRIU.

Could that valve in the RI, from purge ref cell to sample cell, cause pressure variations which might cause RI baseline noise? If so how to fix that? I'm pretty sure this is where the problem is. This morning the baseline was lovely, then I purged and it went back to bad.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
when you purge the ref side of the cell with mobile phase and then run mobile phase in the sample side you get more noise?
17 posts Page 2 of 2

Who is online

In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry