Page 2 of 2

Re: Upper end of concentration range

Posted: Thu Sep 22, 2011 1:00 pm
by Peter Apps
Hi WK

Is it the mls added / 100 mls or the v/v from the density that you use on your x-axis ?

Peter

Re: Upper end of concentration range

Posted: Thu Sep 22, 2011 1:50 pm
by WK
Hi Peter,
From the %v/v determined by tables I convert to %w/w both the ethanol and propan-1-ol for use in the calibration tab in the method.
I then convert samples back to %v/v that have been run in the sequence using factors and the sample density.
WK

Re: Upper end of concentration range

Posted: Thu Sep 22, 2011 2:23 pm
by Peter Apps
Maybe the inaccuracy is in the tables. What if you plot peak areas vs the volume % based on your own volumetric measurements ?.

Peter

Re: Upper end of concentration range

Posted: Fri Sep 23, 2011 7:56 am
by WK
Hi Peter / Yama001
I ran some upper levels last night with the suspected outlier removed and the calibration is much
improved so I will repeat this over the weekend. The control samples run later were all satisfactory.
Thanks for all your help.
Its only by these discussions that I learn !
Regards
WK

Re: Upper end of concentration range

Posted: Fri Sep 23, 2011 7:29 pm
by aldehyde
Hi WK, sounds like you've already tried a lot. Have you attempted the larger volume injection yet? I am interested to see if that helps.. I have seen that the reproducibility of injections at 0.1 or 0.2 uL is a bit worse than at 1 uL. I don't know if it is an effect of the sampler or the syringe. Some people get along just fine with small volume injections (certainly helps keep things clean!) but it all depends on what level of precision you need.

I would investigate the effect of your injection volume if only to rule it out.

Re: Upper end of concentration range

Posted: Sat Sep 24, 2011 7:55 pm
by WK
Hi aldehyde,
I did that initially after Peter suggested it - but it didnt improve my problem which I am certain was the outlier.
It seems I get sufficiently good work with 0.1ul 200:1.
This weekend I have prepared some more levels/controls to spread the calibrations better in each of the 3 ranges.
I will report when I have processed the data.
Best Regards
WK

Re: Upper end of concentration range

Posted: Sun Oct 09, 2011 5:45 am
by mckrause
Flatten out your concentration bias by calculating in log-log space. It will help your top end significantly (as well as your bottom end)