There is a 'reference' button that I can press on the RI. The technicians don't even know how to work these machines, neither does my supervisor really. This is how it runs here at this University. We figure out everything by ourselves which is way I am asking you all for help
hm, ok.
do you have or can you get the manual of your RI? maybe someone of this forum can provide it, if you post the manufacture and model? Or just google for it and/or ask manufacturer.
Basically, a RI detector consists of two cells, one reference and the analytical one. The signal will then be measured as the difference in refraction index of the solvents in these two cells.
Therefore the reference cell has to be filled with your eluent before injecting a sample. During the run, the reference cell is "closed" (excluded from the flow path), that's why RI doesn't work with gradients.
So to fill the reference cell, one has to switch between these two cells either manually or by software (I guess you have to do it manually).
Question: When you push the "Reference Button", is there any inidicator (Display or LED) which indicates the actual selection? hopefully there is... otherwise you have to figure it out how to switch between the two cells yourself.
Without knowing any details of your detector, I would proceed like followed (every time you set up a new mobile phase or start a new sequence) :
- set up your mobile phase, premixed, one channel 100% when using RI, degas it well (if possible by Helium sparging)
- purge pump as usual, then connect the column
-
while equilibrating your column with the eluent (at least 30-40 ml with HILIC on your 4.6x250 column),
select the reference cell (to purge and fill it with the mobile phase).
-
after that, switch to the measuring cell, let it also equilibrate (maybe 5-10 minutes) until the baseline is stable.
- reset the baseline to "zero" with the "Balance bttn". (balance before each injection (if there is no "autozero" function with the software)
- make your injection(s)
Beside this, RI is also sensitve to temperature differences, so it is advisable to keep your column at the same temperature as is your RI-detector. If you can thermostate your RI (maybe there is also a "temp bttn"?), I would use something of 35 or 40°C for both, RI and column. Otherwise I would not heat the column.
Power on your RI detector at least half a day (better overnight) before starting your experiments, so every part is well temperature equilibrated. Leave it powered on if you plan to continue your work the other days or so.
A conc of 10g/L should be sufficent to see some peaks.
Nevertheless, still do the check for flow the rate, so you're sure that there is no other problem hidden.
...and finaly, if you manage to bring your machine to work, write it down in a short SOP. Any of your successors (as well as your supervisor) will be deeply grateful for it