Experimentally is the best way to determine this, rather than doing so mathematically from the regression line data.
The noise level of your GC and the purity of your water and DMAc, the amount of HS sample you inject and the ratio of your split (if you use a splitter) all affect the LOD and LOQ.
You saw from the data I sent you from my paper that my blank values were sometimes zero and sometimes a few hundred counts. One of the problems I encountered was the amount of solvent in the lab air. My lab contained several LCs and the floor of my building had several labs. Thus the building air contained low levels of methanol, acetonitrile, and some of the other solvents. In the spring and fall when the windows were open I would get traffic passing by my building and the exhaust from the shipping and delivery trucks would contaminate the air with hexane and toluene so my blanks for those solvents would not be zero.
And heaven help me if I drank a beer with my pizza at lunchtime !
My ethanol breath would affect the ethanol blanks the rest of the day !
Prepare blank samples with smaller amounts of your std addition solution (compensating the reduced volume with pure DMAc) until you are only adding 10µL of pure DMAc. When you reach the lowest level of solvent in which you can see the solvent peaks then you have reached your LOD and usually about 3x that level is your LOQ. Be sure to check that the addition of your drug to that concentration of solvent does not affect the LOD, or at least you measure how much the drug affects that value, either higher or lower.
best wishes,
Rod