The ultimates i know draw the sample into the loop by the syringe and the put the whole loop in the flow path. So you will get solvent int the syringe. don't think this is a good idea as 50/50 Methanol/water and 50/50 cyclohexane/n-heptane don't mix. take a close look at the syringe.Autosampler is an ultimate. Do you mean, the rinse for the autosampler syringe needs to similar to the mobile phase, because currently I've been using 50/50 Methanol/water for all methods.
- is the syringe solvent degassed?
- any bubbles in the syringe?
- do you puncture the vials, (500 µl is a lot)?
- is there enough sample in the vial (sounds stupid, however, we had issues other months with very very experienced stuff)?
- are the connections between needle and loop tight?
Good luck,
Alex
(Who is enjoing precise 1000µl-injections on a ultimate sampler)