I'm working on wines, liquors, honey and soft drink sugars content.
Flow: 1 ml/min; eluent 80:20 ACN/ Water plus 0.2% Triethylamine. Samples at 25°C Column 35°C RI detector 35°C
I've destroyed two column because the increasing pressure.
Now, at the end of the day I clean up the column as follows:
First change to 80:20 ACN/water without Tryethylamine,
Then change to 50:50 60 min
then change to 50:50 ACN/Water plus 0.1% Trifluoroacetic acid and hold 60 min
New change to 50:50 ACN/Water 30 min abd finally store the column in 80:20 ACN/Water
I'm working like this three months and pressure increase has been no significant (Initially 1175psi, nowadays 1250 psi)
The column is 250 x 4.6 mm
Perhaps it can help
Guillermo