You can try to isolate the column from the FID and see if the signal is still high, then you can find out the reason. The quality of compressed air is also one of the reasons. You may have to put in a zero air filter.
So you have some long term drift in standing signal. Does this affect your analysis ? Have you done any of the troubleshooting that has been suggested ?

Peter
Peter Apps
So you have some long term drift in standing signal. Does this affect your analysis ? Have you done any of the troubleshooting that has been suggested ?

Peter
I've changed some traps but had no chance to check result yet. I don't know how it can affect my analysis.
All I ever need to know I'm learned in cloning vats.
The first thing that you need to do is to remove the column from the detector, plug the connection, and check the baseline and whether it drifts. This will immediately tell you whether the carrier gas is part of the problem.

Unless you know that this is a problem with an actual analysis you might well be worrying abut nothing.

Peter
Peter Apps
What are your flow settings?

If you have just air and hydrogen (no make up) does the signal drop? I have seen contaminated gases do exactly this.

If you have contamination in air/hydrogen lines it may be difficult to tell if there is an effect.

Is this plumbed directly between gas tanks and GC or is there common gas supply to other working instrument?
Well, It is revealed, that traps had not been replased for years at this GC. I'm going to check all of them. (moisture trap in H2-generator and Hydrocarbon trap at air flow are firts at a row).
My bad, I should be more accurate with instrument's past next time.
Thanx for your assistance!!!
All I ever need to know I'm learned in cloning vats.
Once a trap is saturated with hydrocarbons/the oxygen scrubber is depleted it turns into a sponge that just emits constant levels of contamination rather than helping you :).

A very common problem.
A very common problem.
Yeah! I'm in process of recovering H2O trap from moisture now. Still I want to place air generator away from HPLC's. But I don't think you can help me in conversation with my boss :wink:
All I ever need to know I'm learned in cloning vats.
23 posts Page 2 of 2

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry