Re: transfer line from GC column to MS
Posted: Wed Feb 02, 2022 8:36 pm
MSCHemist, it sounds like we have something similar to you in terms of this quick swap/backflush set up. Is yours an Agilent system?
A couple of years ago a service engineer from Agilent came by and told us that with experience they had found that having 4 psi He on the line isn’t the way to go. I think they had said 4 psi as a standard when setting up these quick swap fittings initially.
If this is relevant, you could contact Agilent or contact me directly and I will try to remember all the technicalities as to why. We made the changes they proposed and things more reliable
Kevin.