Re: GC- Calibration for Nitrous-Oxide- diluting gas
Posted: Mon Jun 06, 2011 12:55 pm
ashrestha,
Since you have access to 100% nitrous oxide, I would recommend using the dilution scheme set out by Rodney. Heise makes gauges which can be used for this sort of work. If it were me, I would dilute down to what I thought was 100 ppm and run that against the other mix I bought at 100 ppm.
Best regards,
AICMM
Since you have access to 100% nitrous oxide, I would recommend using the dilution scheme set out by Rodney. Heise makes gauges which can be used for this sort of work. If it were me, I would dilute down to what I thought was 100 ppm and run that against the other mix I bought at 100 ppm.