by
punt » Wed Oct 05, 2011 3:49 pm
Hi Rod,
After much troubleshooting, i realised it could be the moisture that cause my H2 peak shape to change from fronting to symmertical. When i inject a lower concentration of H2 like 60%, i can see that the peak is tailing. After i condition the Molsieve5A column at 280C for few hours, the peak profile change from symmetrical back to fronting and the area and retention time is now more stable. Thanks for your useful information.
Some question i would like to ask you, for molsieve 5A and 13X column, how many times can we regenerate at high temperature? Usually wat kind of symptons do we see for such column before we decide to change to a new column? Other than moisture, how does CO2 absorption to the molsieve column affect the column performance? Thank you
AICMM,
I did not try to change to a smaller sample loop becoz i dun think the problem lies with the overloading as i have 2 similar GC with the same configuration. If its a overloading issue, the other GC which is alright should be facing the same problem. Moreover when i Inject a 60 percent H2 Std, the peak shape also look fronting.
Thank you
Regards
Punt