1. It is definitely better to have the same %recovery for both the IS and the analyte (i.e. "relative recovery" of 100%).
2. However, if you carry your calibrator solutions through the same workup procedure as your samples, and if your relative recovery is in fact constant (e.g., always 90%), then you should get the same relative recovery in both calibrators and samples.
3. In that case, the relative recovery effects will cancel out, and you will get the correct values for your analyte in your samples.