Re: Experiencing poor Sucrose elution
Posted: Wed Apr 13, 2011 9:52 pm
Hi
With your existing equipment and your separation needs, I would recommend to use an amine column - some similar suggestions have alreay been made. The new Sugar-D column looks quite interesting but I have no experience with it. I normally use the Asahipak NH2P-504E (from Shodex), but they have just released a new version of this column with smaller particles which looks interesting. The only possible issue you might have with these types of columns are the trisaccharides and larger. You need to use isocratic conditions with the RI typically a mixture somewhere in the region of 80/20 to 70/30 AcN /H2O will resolve the mono and disaccharides. I think the longer oligos are eluted with a bit more water. The amide columns (e.g. Tosoh Amide-80) should also be able to perform the separation under similar conditions - but again I don't know about the situation with the trisaccharides.
If you had an ELSD, or CAD then the above solutions would be fine since you could introduce a gradient. Another option that would allow you to use a gradient would be to introduce a post-column derivatisation that would give you something detectable in UV/VIS - there are several options described in the literature for this, but you would need an additional pump for introducung the reactants.
Pre-column derivatisation would also be a good solution. But most of these derivatives exploit the reducing end of the sugars for labelling - so it won't work for sucrose. I don't know if a suitable label exists that you could use for sucrose.
Hope some of this is helpful.
Sean
With your existing equipment and your separation needs, I would recommend to use an amine column - some similar suggestions have alreay been made. The new Sugar-D column looks quite interesting but I have no experience with it. I normally use the Asahipak NH2P-504E (from Shodex), but they have just released a new version of this column with smaller particles which looks interesting. The only possible issue you might have with these types of columns are the trisaccharides and larger. You need to use isocratic conditions with the RI typically a mixture somewhere in the region of 80/20 to 70/30 AcN /H2O will resolve the mono and disaccharides. I think the longer oligos are eluted with a bit more water. The amide columns (e.g. Tosoh Amide-80) should also be able to perform the separation under similar conditions - but again I don't know about the situation with the trisaccharides.
If you had an ELSD, or CAD then the above solutions would be fine since you could introduce a gradient. Another option that would allow you to use a gradient would be to introduce a post-column derivatisation that would give you something detectable in UV/VIS - there are several options described in the literature for this, but you would need an additional pump for introducung the reactants.
Pre-column derivatisation would also be a good solution. But most of these derivatives exploit the reducing end of the sugars for labelling - so it won't work for sucrose. I don't know if a suitable label exists that you could use for sucrose.
Hope some of this is helpful.
Sean