Re: Recurring ghost peak
Posted: Fri Apr 08, 2011 6:34 am
Sometime there was a 'bleeding contamination' which fouled the pneumatics upstream from the injector. It may have come from the gas supply originally, or from the pneumatics hardware (grease, oil, lube of some sort) or from sample backflash from the injector. But when a spot in the gas flow is cool enough and there is a phase for the contamination to be trapped and focused into a plug, when the column is heated it proceeds through the column to the detector as a 'peak'. With more time it becomes larger. If the column is not allowed to cool, it elutes as a 'smear' or baseline increase hidden within the 'bleed' of the column.
best wishes,
Rod