Recurring ghost peak

Discussions about GC and other "gas phase" separation techniques.

17 posts Page 2 of 2
Sometime there was a 'bleeding contamination' which fouled the pneumatics upstream from the injector. It may have come from the gas supply originally, or from the pneumatics hardware (grease, oil, lube of some sort) or from sample backflash from the injector. But when a spot in the gas flow is cool enough and there is a phase for the contamination to be trapped and focused into a plug, when the column is heated it proceeds through the column to the detector as a 'peak'. With more time it becomes larger. If the column is not allowed to cool, it elutes as a 'smear' or baseline increase hidden within the 'bleed' of the column.

best wishes,

Rod
The old regulator has almost certaialy contaminated all the gas plumbing, so the blank peak is still bleeding onto the head of the column. There is very likely a small filter immediately upstream of the inlet - change that and then let the gas flow overnight (with a decent split flow at the inlet so that most of what bleeds off the last piece of tubing does not get oto the column). Then do a few runs.

Peter
Peter Apps
17 posts Page 2 of 2

Who is online

In total there are 165 users online :: 0 registered, 0 hidden and 165 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 165 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry