Baseline problem or something ...

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

27 posts Page 2 of 2

syx,

I am having a problem with baseline (sudden drop during a gradient run). I would like to post it on this forum but it's hard to describe the baseline without a picture. could you let me know how you got the picture in your posting? Thank you in advance!

Regards,

JZT

Ok, seeing the chrom now: It really looks like what I have seen, sometimes, when air was co-injected, though I don´t understand why your blank doesn´t do this (why don´t you inject H2O?). Do you concentrate the samples with a stream of gas?
I will try to inject H2O as you suggest. I think the blank is mobile phase or sample diluent, and I didn't use water as the diluent.
I do not concentrate samples.
I would like to post it on this forum but it's hard to describe the baseline without a picture. could you let me know how you got the picture in your posting?
Jzt, open <http://imageshack.us/>, register, hosting an image, and copy-paste the code they give to this forum.

The "active substance solution" contains other stuff than just your active ingredients. A polymer, maybe...?

hmmm... I don't think it has a polymer inside the bulk. I will try pH effect next week. I will send the report here.

Is there anything else in your active substance solution? Some polymeric ingredient?

... maybe its impurities. Can it be a polymer?

I have tried pH effect to separation. In lower pH (I tried pH 4.5 and 2.5) the main peak will loose its retention with shallow tail.
The problem is reduced when I try to inject lower concentration of the substance.

:?: What diluent are you using for Sample preparation?

I use mobile phase.

Thank you Mr. Tom.

I try to use higher pH value of mobile phase.
The pKa value of the substance is 8.91.
The maximum pH can be used for LC system and the column is 10.
Based on those 2 facts, I decided to use ammonium buffer adjusted to pH 9.0 - 9.5 and found three peaks with no baseline problem.
First peak is related to degradation product of the main substance in basic circumstance. So,. I use the mixture of water – acetonitrile instead the mobile phase, but still in the same strength.
The last peak (the smallest) is unknown. It has same spectra with the main peak. Is it probably the enantiomer?

Ok, here is my wild guess. You have a degradant or impurity in your TEA that is normally at equilibrium. Your injection is disrupting this equilibrium and what you are seeing is the system returing to equilibrium.

What is the size of your injection? Have you tried without TEA, etc. I would try switching out your reagents individually with fresh high purity ones and try to locate the problem. Good luck and let us know how it turns out.

What is the size of your injection? Have you tried without TEA, etc. I would try switching out your reagents individually with fresh high purity ones and try to locate the problem. Good luck and let us know how it turns out.
Injection volume: 20 uL.
If we did not use TEA in lower pH of mobile phase (as we use in first post), the peak would be shallow with long tail.
In pH 9 or greater, we do not need TEA to get a sharp-thin peak.

Image

Condition:
Column: 250 x 4.6 mm, contains 5-um packing L1
Temperature: 25 C
Mobile phase: mixture of acetonitrile and 0.05M ammonium acetate buffer pH 9.0 (55:45)
Main peak is eluted in 27.2 minutes after injection.
Asymetry: 1.2; Theoretical plate over 9000. Resolution of the last peak (Rt 30') to the main peak is 2.38.
27 posts Page 2 of 2

Who is online

In total there are 8 users online :: 2 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry